Hi Taylor,
With Med Associates equipment, definitely make sure that the output signal has the polarity and voltage range you need to trigger Pulse Pal.
On an old Med Associates assay, I remember that the standard logic signal was something like -28V with respect to signal ground. If this is true for your system, it presents two issues:
- 1.
The voltage is above Pulse Pal's trigger channel range of +3V to +5V (see Pulse Pal specs here). Over time, using a higher voltage can damage the optoisolator.
- 2.
The voltage is negative, and needs to be positive
The first issue can be solved by creating a voltage divider with two resistors (e.g. this).
The second can be solved by simply reversing the leads. This works because Pulse Pal's trigger channels are optically isolated.
If you're seeing the trigger channel LED light up, you have the correct polarity and Pulse Pal has received the trigger signal. I'd check the trigger link parameters of your output channels, to make sure they're configured to respond to trigger channel 1.
Please let us know if this helps!
-Josh